Sensor test Diverite Liberty, which deviation is acceptable?

Please register or login

Welcome to ScubaBoard, the world's largest scuba diving community. Registration is not required to read the forums, but we encourage you to join. Joining has its benefits and enables you to participate in the discussions.

Benefits of registering include

  • Ability to post and comment on topics and discussions.
  • A Free photo gallery to share your dive photos with the world.
  • You can make this box go away

Joining is quick and easy. Log in or Register now!

Agro

Contributor
Messages
841
Reaction score
185
Location
Zurich, Switzerland
I have not beeing diving the Liberty for 4 month. Today I made the sensor pressure test, 1 to 3.5 bar pressure. Liberty shows deviation of ideal numbers, I had a max of 8% at 3.5 bar.

My old manual sais 7% is the max allowed deviation. Is this still Divesoft's recommendation? I found no information on Divesoft's website. Am I blind?

As far as I know sensors "go to sleep" if they have not been used for a long time. I must wake them up by exposing to oxygen for some time. I guess a sensor test at 3.5 bar is a perfect wake up. But can I trust this first test or must it be repeated?

I thinks I will keep the sensors at 1 bar of oxygen for 24 hours and then repeat the sensor test. What do you think about it?

Sensors are 8 month old.
 
Some great reading (attached PDF) here, might be helpful re deviation:

 
Some great reading (attached PDF) here, might be helpful re deviation:

Well, some interessting points but unfortunately not the answers to my questions.
 
Im very new to RB but why would you test up to 3.5 bar? Any deviation past approx 2.0 bar is meaningless. If your sensors are linear to 1.6 bar you are good to go.

Keeping them exposed to 1 bar to 'wake them up' will kill the life span also. Mine 'woke up' by air exposure for a few days.
 
Liberty shows deviation of ideal numbers, I had a max of 8% at 3.5 bar.
Interesting that the test goes this far out of the working range. Given that the highest you typically dive is 1.6, I would assume that non linearity beyond 2.0 is not so relevant. How are the 3.5 readings used? Just compared to an expected deviation or used to predict deviation in the working range of 0.5-1.6?
 
1 bar pO2=50mV
3 bar pO2 should be 150mV
In reality 3 bar pO2 is 135mV, so deviation is 10%
Reference is always measurement at 1 bar

A new sensor is linear up to 3.5 bar. After for ex. 18 month it is linear up to 2 bar and not linear anymore at 3 bar. After for ex. 20 month it is linear up to 1.5 bar and not linear at 2 bar.

Testing at 3.5 bar gives you a good information about sensor's shape.
 
Testing at 3.5 bar gives you a good information about sensor's shape.
If the cell is linear at 2 and non-linear at 3 bar that cell is fine. Testing way up at 3.5 bar is not useful information.
 
No, that's wrong.
Every galvanic cell ever made will lose linearity at 3.5 at some time, some won't even get there new out of the box. That tells you nothing about it's fitness for use below 2.0.
 