rmssetc
Contributor
You know, I just became aware of another failure scenario which could be rather unpleasant to a simple rec OC diver: the cell is current limited below 0.21, but the multiplier circuit can bring the voltage up to showing 0.21. But then the output would be flat from below 0.21 and beyond my max fO2, and my tank of properly mixed EAN36 would be identified as a tank of air. Could be nasty if I'm planning to go to 40m...
Fortunately, most of the unpleasant failure modes usually will be detected by using two different analyzers. And if they don't agree within the expected variance of ~1 percentage point, don't breathe the tank until it's been analyzed with a third unit which agrees with one of the others.
While accurate, that's a pretty amusing suggestion, considering that the original thread (Nitrox analyzer: to take or not to take?) had a lot of responses about the local analyzer being unavailable (missing, busted, busy). If you don't trust one analyzer, what are the odds of 3 being available? Maybe the question should be "Nitrox analyzers: to take three or not?"