aue-mike
Contributor
Mike,
I believe you are thinking that as the diver descends on CCR, the loop PO2 will normally increase and can even spike high during a rapid descent typical of technical dives. However, my statement is not at all "problematic". I've investigated several accidents and personally observed one where LOC occurred during or shortly after descent. While descent does cause a relative increase in the PPO2, at the same time the diver is burning oxygen out of the loop that is not being replaced if oxygen supply has failed.
Here is a general example, one of many possible: The CCR diver has properly prepped their unit including a pre-rebreathe during the transit to the dive site. During the transit, the oxygen valve is 'rolled off' as it bounces against a tank rack and bench designed for AL-80's. The loop PO2 is stable at the expected set point, say 0.7 ATA from the pre-prebreathe and indicated pressure on the SPG is as expected as well (from the charged HP hose, which is quite long on some units and holds a surprising amount of oxygen).
Once the dive boat reaches the site the diver dons the CCR, runs an abbreviated pre-jump list, then enters the water and makes a very rapid descent. In our area, this is a typical occurrence especially if the diver is "smart bombing" the wreck. While the loop PO2 is initially increasing, and the solenoid initially injects some oxygen (from the hose residual), eventually the loop PO2 will begin to drop and will rapidly drop if the diver pauses their descent for any reason. The first diver jumping on a technical wreck dive in our area often has the task of chaining in a descent/ascent line for others to follow. That diver is exceptionally task loaded with navigating to the wreck and tying in. If the task loaded diver fails to notice their setpoint is not being maintained (the usual concern being a brief hyperoxic spike, not hypoxia), then LOC follows.
Note, this isn't a hypothetical example, it's has happened to me personally exactly as I described and I was lucky enough to notice dropping PO2 before the loop became seriously hypoxic. I've caught two more roll offs over the years, but before gearing up as I now always manually check the valve immediately before gearing up. I've also personally observed several CCR divers actually perform their pre-breathe and then close the valve as a result of an OC habit. Just to be clear, I'm not proposing the circumstance I've just described above as what happened in the case of the accident being discussed. I'm simply describing my findings in more than one case where LOC on CCR happened during or very shortly after rapid descent.
Appreciate the response Mark. Yes, these scenarios can happen as described. And I am aware of incidents where divers have gone hypoxic and perished on the bottom due to O2 valves being turned off. But IMHO this speaks of poor training, experience, and practices.
EDIT:
"While the loop PO2 is initially increasing, and the solenoid initially injects some oxygen (from the hose residual), eventually the loop PO2 will begin to drop and will rapidly drop if the diver pauses their descent for any reason."
I don't think I would agree with your assertion that the PO2 would "rapidly drop if the diver pauses their descent for any reason." During our CCR training, our instructor had us set our PO2 at a safe level (can't remember the specifics, but let's say, e.g., 0.50) and then shut off our O2 to monitor the rate of our PO2 drop while our instructor observed/monitored the situation. During our course, we even kicked against the side of the pool to see how fast one could burn (metabolize) through O2. It was definitely not "rapid" in our experience. But perhaps I am misunderstanding your point.
I know we could easily delve into semantics and "chicken versus egg" discussions, and will concede these scenarios would ultimately be considered hypoxic fatalities, but that is not, nor should not, be the real trigger. It's clearly human error.