For a rebreather, having a cell that reads above one atmosphere partial pressure is a must.
Not sure why you need that overpressure for an analyzer?
1. A good cell has linear reading from 0 to 3 bar.
2. An old cell often is linear from 0 to 1 bar, non linear from 1 to 3 bar.
3. A very old cell is non linear from 0 to 1 bar.
If you test a cell from 0 to 1 bar you only detect situation 3, not situation 2.
No 2 tells me, that No 3 will happen within the next time.
If I test my analyser cell at 3 bar and get situation 1 I know cell is good and will probably still be good from 0 to 1 bar within the next time.
If I test my cell at 3 bar and get situation 2 I know cell is good from 0 to 1 bar but may fail within the next days.
If I have a cell which sometimes gives good results, sometimes bad results, the cell might be between situation 2 and 3. By testing it at 1 bar I might get result 2, then I believe everything is fine. By testing it at 3 bar I will detect the problem.
I my opinion it is a good idea testing a cell at 3 bar even if it is used in an analyer only. But I am very open to other opinions.