You did not list the mV @ 1.4 for this cell so I can't give the exact answer because we don't know how the cell changes between our datapoints from 10% to 15% deviation. However if I was diving 1.6 setpoint on the computer the actual ppO2 in my breathing gas would be 1.7. The computer is looking for 68.8 mV for 1.6 but we tested that the cell only outputs 65mV at 1.6......If you are 10% off at 1.0 and diving 1.4 as a set point you could easily be at or above 1.6 simply due to lack of linearity.....
Slacker,
43mv is the actual 02 output which gives a 10% linear deviation. "If" the cell follows the same linear deviation then it will continue to deviate past 1.0. If the PO2 is bumped up to 1.6, for decompression, then the actual set point would be 1.84 when reading 1.6. Linear deviation is not normally a straight line from linear it is normally a curve. I was attempting to keep it basic and simple so that it would be easier to follow. I also need to spend some time putting in graphs that properly visualize how linear deviation normally works. I believe that working from 0.21 up and showing a normal curve from that point is the easiest for most to understand however there could be a better way to present it, which I'm very open to suggestions.
I'm referring specifically to the example in your pdf. (cell reads 10mV in air, 43mV at 1.0, and 65mV at 1.6) where you correctly state 15% linear drift (@1.6 reference) from air. You then state with a setpoint of 1.4 you could easily be above 1.6. I'm saying the math there is incorrect and your will not be above 1.6.Slacker,
I'm not saying that there is 15% deviation from air, I'm saying that there is 15% deviation from O2 @ 1.0. 1.6X1.15=1.84. You would be seeing 1.6 however actually be diving 1.84. This is with a correlated linear curve which easily might not be the case, reality is that you could be over 2.0 with no way of knowing. This is what we call limiting however it is really linear drift that we can't see because of the human limit on PPO2 tolerance.
I apologize if I'm not stating this clearly.
Using O2 as calibration standard will give you higher accuracy than using air but only when your sensor is new. However, since the sensor's linearity is lost starting with the higher O2 values, using O2 as a standard will give you higher error than using air if your sensor is aged. And this will always work one way, giving you values lower than the real numbers. Also, using O2 for calibrations will age your sensor sooner.Bobby,
I understand that linear deviation is not a straight line which is why I used the values you picked in my example. Baselining on 0.21 may be easy to present but it is not the way our computers work. I calibrate against O2 not air. If I have calibrated my cell in O2 at 43mV my DC shows 1.0 at 43 mV. It doesn't care that it was 10% non-linear before or after that. I have changed the scale to reflect the 43mV reading. So your example that showed 10% deviation from 0.21 to 1.0 and 15% deviation from .21 to 1.6. So using air as your reference you say the setpoint is 1.84 when reading 1.6, but the computer would actually be showing 1.67 because it was calibrated @43mV and now reading 65mV.
Wanting 1.6 and getting 1.84 could be a big deal, but getting 1.67 instead is less of an issue.
Bobby, from the paper, it appears that you are not aware what is causing sensor aging and how exactly the linearity is lost. If you knew, you would not have said that "Voting logic simply shows, or votes out, a cell that has reached enough deviation from the other two cells. It does not determine which cells are correct." Because if you have 3 analyzers calibrated to 21% with the same air tank side by side, and then you check the same Nitrox tank using these, the highest reading will be the most accurate. Your cell with a 10% deviation at 100% will most likely have only a 1% deviation at 32-36%.Dave,
Please let me know what isn't correct in what I wrote, either through a post or PM.
Thanks,
Bobby
I'm referring specifically to the example in your pdf. (cell reads 10mV in air, 43mV at 1.0, and 65mV at 1.6) where you correctly state 15% linear drift (@1.6 reference) from air. You then state with a setpoint of 1.4 you could easily be above 1.6. I'm saying the math there is incorrect and your will not be above 1.6.
Again, please remember the paper is a draft that was done years ago and hasn't been formally reviewed