madmole once bubbled...
In English (as opposed to American), common use for "Regulator" means to fix a value. ie a speed regulator on a steam engine, or a pressure regulator on the Central heating (that vents above a certain pressure) or the Flow regulator on your HP hose.
Ah, another Americanism. I had to go and look him up.
Now I'm arguing just for the sake of arguing
Unfortunately their are some failure modes in which more than one cell could fail. A good example would be 2 old cells that no longer have sufficient electrolite to reach the output required for 1.3 bar of PP02. in which case they would calibrate fine at the surface, but at a depth of greater than 3 m would not be able to respond to a 1.3 bar PO2 (they would indicate a lower value) as 2 of the cells are in rough agreement the 3rd good cell would be outvoted and ignored resulting in continous injection of O2 leading onto an O2 convultion. There would be no warning alarm or indication in this case. However this is easily prevented by
a) replacing the cells at yearly intervals (they are rated to 18-24 months constatnt use)
b) listening to the solenoid firing (you get a good feel for how often it should fire, once every 6 seconds would be very suspicious)
c) All inspiration users should do a quick test apon reaching the bottom, you inject O2 to take the PPO2 to 1.6 bar quickly, this verifies all cells can respond fully in the range we need
d) You can do a diluent flush at any time to bring the O2 to a known value regarless of electronics.
roakey once bubbled...
Since the claim has already been made that George Irvine is a bigot, I guess it won't hurt to post some recent mail from DIR/Quest where he gives the inspiration both barrels.
Might explain why the deaths/unit rate on the thing is so astronomically high versus OC, ESPECIALLY when you limit the sample to experienced divers.
Roak
=====================================
Wrom: JMVRESKPNKMBIPBARHDMNNSKVFVWRKJVZCMHVIBG
To: "Quest@Gue. Com" <quest@gue.com>
Subject: Desperation Rebreather - shovel and all
Date: Thu, 15 Aug 2002 16:33:46 -0400
Since those who do not know want to make an issue of me not being willing to
waste my time talking about why elephants can't fly, I will waste all of
your time with this one to get at those who keep recommending the Buddy
Inspiration, and who apparently have an agenda with this thing and other
shade tree rebreathers out there.
Besides all the failure modes common to all rebreathers, like if the
mushrooms in the mouthpiece malfunction and take the diver hypercapnic or
hypoxic right away, or a loop hose failure, etc, the Desperation has its own
plethora of problems idiosyncratic to the unit. Other units have many of the
same issues, and almost all of the civilian rebreathers are poor excuses for
anything other than assisted suicide devices. In fact, most are merely
copies of each others bad ideas, and most are poorly funded , poorly
capitalized, poorly tested and otherwise poorly conceived wet dreams. The
Halcyon is a clever device, made more clever by Reinhard Buchaly, but is
still a dangerous piece of "kit", whereas the Buddy is a dangerous piece of
"****".
I will talk about the Inspiration Rebreather as it comes from the factory.
There are many ways to modify the rebreather as one may observe on the Star
Wars Bar Scene lists like "rebreather@nwdesigns.com" and is a complete topic
onto itself. I will stick to
the basics here and leave the "fix it at home" mods to the die hards on the
other lists.
There is a small metal screw inside the oxygen sensor compartment, it is
located directly across from sensor number two. Condensation often forms on
this metal screw and will allow droplets to fall onto the oxygen sensors.
This will happen particularly if the diver moves from side to side as in
dumping gas from a drysuit or if they invert from horizontal for any
reason.
Dont know about you George, but I'd definately want to be measureing the Oxygen level somewhere between where it was mixed and me breathing it. Where else are you going to put it ? (up your arse hopefully as thats where all the hot gas is) . have you ever seen an actual unit. The sensor have PTFE coatings so water CAN NOT sit on them, it just runs off. At least we have PPO2 monitoring unlike some unitsCondensation is inherently found within this area and will form on the
oxygen sensors even without this metal catalyst. The O2 sensors are located
on the inhalation side of the breathing loop, so you have warm gas that
just
went through the scrubbing process meeting with cooler gas that you will
inhale thus the condensation forming on the cell faces. This condensation
causes discrepancies/inaccuracies within the cell readings and they begin
to
VOTE trying to figure out which one is more than .2 bar out of line with
the
other. Cell warnings will manifest within this period of time and the diver
will begin to get audible and visual alarms ...task loading increases.
If I want 100% O2 I'd use an OC reg on the thing. This shows GI doesn't understand rebreathers. having been on a very high PPO2 for the WHOLE, and LONG dive, most people with brain cells dont actually want to be on 100% as they are at this point near their CNS limits. 80-90% is high enoughThe unit alarms if it senses a PO2 over 1.6, which is a good thing. Problem
with this is that many of the divers will run 100% oxygen at 20 ft which is
a PO2 of 1.6, if they drop below the 20 ft they get an alarm, fair enough.
If you have several Inspiration divers in close proximity with cell warning
alarms, and high PO2 alarms it becomes very difficult to know if the alarm
is coming from your unit or from another diver. Some will be able to
assimilate this to being in an area where several cellular phones begin to
ring and everyone pulls out their phone to see if it was theirs. Mix this
with CCR divers using wrist computers that alarm and you really have an
orchestra playing down there, so much for the peace and tranquility of "no
bubbles".
No must about it. Different divers prefer different attitudes. if you dive with minimum loop vol (as you should) theis isi not a problem. Hey George, how does the Halcyon unit store its gas then if not in a counterlungWeight must be placed on the top of the rebreather in order to balance the
trim. If divers put to much gas in the counter lungs the upper body is
lifted and trim is then off center. Gas in a counter lung is just another
source to administrate, along with drysuit, BCD, PO2 on handsets, pressure
in O2 and DIL cylinders.
yep, 3 hours at 20m with no stops!!!Many divers use the Inspiration to extend their times in open ocean, this
Let me see air in dry suit and counterlung on CCR and Drysuit and BC on OC. Why is this different. And this task loading from a God like you who can manage 2 scooters and 20 stage cylinders at oncein
itself may pose problems. If the decompression gained by the increased
bottom times is met with undesirable conditions such as rough seas it is
not
so easy to adjust buoyancy as with breathing open circuit. Many of the
buoyancy characteristics involved with rebreathers require a longer
learning
curve and must be anticipated, if the diver is not up to par they are much
less forgiving than OC. If the constant PO2 is increased or decreased too
quickly due to unforeseen circumstances the diver could quickly become
hypoxic/hyperoxic.
I can fit one of the built in mouthpieces if I want to. But why would I. George, you have limited your bailout options. How do you swap mixes on OC bailout is they all have to come through the one DV? I can carry bail out or borrow from my buddy. They all have their own DV'sMouthpiece does not have OC bailout built into it, bailout is a time of
increased stress so it is pertinent the transition should be smooth without
chance for a mistake. The bailout procedure on the factory unit uses a
device called an Auto Air, this duals as a breathing device and BCD
inflator. I couldn't imagine this being an effective tool for gas sharing,
nor proper bail out for the user either, as the CCR mouthpiece has to be
effectively closed before the transition to OC bailout is performed or it
will flood the breathing loop making the diver negatively buoyant. Bail out
should be in the form of a combination unit on the mouthpiece to facilitate
safe transition.
Thats why you clean it and check it before each dive. ALL dive equipment has O rings that it would be nasty for them to leak, including your beloved halcyonIf the O-ring on top of the cartridge lid is dirty or not aligned properly
CO2 will take the path of least resistance and bypass the carbon dioxide
scrubber therefore breathed back into the loop. Hypercapnia begins and the
diver is faced with another problem to solve.
Funny that a British woman has just been to 520ft without an ADv and her lungs didn't "Collapse" maybe you should attend some basic scuba lessons. Also you can feel the squeez easily. the task loading is only that for the first couple of dives, before you master itAs the diver descends they must equalize the counter lungs, if this
procedure is not adhered to and they begin an uncontrolled descent the
lungs
collapse and the diver is not able to breath, an automatic diluent add is
an
aftermarket product which does combat this. But since we are talking
factory
here the diver is faced with equalizing counter lungs, ears, sinuses, mask
drysuit, BCD, monitoring PO2 on handsets, buddy position, light and depth
in
the water column. It has been mentioned before that this is "a busy time".
Agreed on the backplate. The fittings however are usual AP vlaves ones that easily stand up to the rigours of diving (and UK cave diving, which is far more hostile thanyour great holes). As for the clips. Mine are 2 years old and fine. A spare comes with the units, but they are actually a LOT stronger than they look.Inspiration does not have SS backplate and utilizes many plastic fastex
clips, which some view as failure points. There are seven quick releases on
the soft harness including the crotch strap and handset clips. The clips
that hold the yellow casing lid on the unit break frequently so spares are
required as well.
Same for all scrubber material in all rebreather. Most of us carry our own spares. What we want to know is if we can buy Scrubber there rather than have to transport it and can we get an O2 fill. You wouldn't know about such things, as you only dive your nice comfy Waukala hole. the rest of us like to dive in different sitesOften difficult to transport with Sofnolime, if you do not have an MSDS on
your person you will be declined, even if you do have the Material Safety
Data Sheet on hand and the handler does not feel comfortable with the issue
they will not let the scrubber material on the plane. Most Inspiration
divers seek out "Inspiration friendly dive Centers" so they are able to
obtain the wide array of parts required to service and maintain.
You can put whatever sensors you like in the unit. The warranty and CE approval is only valid if you use the correct ones. Doesn't the halcyon manufacturer sell cells. Oh, I forgot, its never undergone any third party testing or approval. The Inspiration meets all European safety standards for Personal Protective Equipment, including breathing resistance, reliabilty, Scrubber duration and breakthrough, and has been independantly tested to gain thisThe oxygen sensors are proprietary to the Inspiration which limits the
diversity on this product. Many of the CCR's will allow various types of
sensors to be used but not so the case with Inspiration's and it is
strongly
voiced by Martin Parker. Patrick Duffy with Oxycheq in the US sells similar
sensors and says there is absolutely no difference between the Teledyne's
he
sells and the Inspiration sensors.
Scrubber duration is what it says it is, 3 hours. this has been tested by offical agencies under high CO2 output and flow conditions. No other rb has been tested this way. If you choose to extend the duration, thats your life. Has the halcyon been tested,like this, No, its duration is based on the "GI got 5 hours out of it last week" methodScrubber canister is small (2.45 Kg of 8-12 mesh, 797 diving grade
sofnolime) and does not facilitate the use many of the mixed gas
Inspiration divers put it through. At depth CO2 breakthrough is rapid even
with a resting diver, if breathing resistance is elevated the scrubber is
near void. With increased CO2 build up the diver is of course exposing
oneself to further malady. Diving high helium concentrations assist with
this problem as it is less dense than air, easier to breath therefore less
CO2 buildup and the scrubber should last longer but it is playing on the
edge. High PPN2 should be ultimately avoided.
A 1 or 2 incgh movement will make SOD all difference. But has anyone broken an AP fitting??? yes iyt does have lots of quick releases, so that when you are hanging on the side of a UK dive boat in a swell and current, trying to get up a swinging ladder you can dump it quick and pull it in afterwards. Something the DIR, one piece straight jacket people cant doIf counter lungs are not situated adequately they will float above the
divers shoulders and increase breathing resistance. It is taught in the
basic course to watch for this ... But they are clipped down with fastex
buckles which as we know do fail on occasion. With the diver already quite
task loaded on the CCR it is easy not to notice the lung has crept up, CO2
will then increase from breathing resistance.
ytes there is and your about to give itThe LP hoses which feed the diluent and the oxygen inlets on the counter
lungs use a different end than the BCD inflator. The BCD is inflated with
Diluent gas, the same which you are adding into the counter lung, there is
no sound reason not to have the same end on this for diversity.
crap, a few very vocal ones do. The vast majority continue to use it. Personal preference, oh I forgot you are the kit Police and we MUST wear what you approve. GI you have NO experiance of UK open water conditions and Caves at all. I dont tell you what to wear in your gin palace caverns and you shouldn't tell me what to wear in conditions youve never experianced. the DIR kit is NOT suitable at all in many of them and would get you killed very quickly in most UK cavesThe reason
the end is different on the BCD inflator is to supply a greater amount of
gas to the Auto Air regulator used as a bailout/inflation device. This Auto
Air is prone to free flow situations and can dump the diluent gas if not
tended to quick enough. Most Inspirations divers discard this Auto Air
early
into their CCR career.
Why would you shut off both handsets. Do I worry about you turning off both of your twinsets, no, you wouldn't do it, neither will I turn off both handsets.If both handsets shut off in the water the diver is faced with a series of
questions in order to "reboot" the system. One of the questions ask if you
would like to calibrate "yes or no" if the diver is stressed and chooses
"yes" they will effectively be adding 100% oxygen into the breathing loop
no
matter what depth they are at in the water column.
if the battery is low, it gives a low battery warning and you can swap the handsets over. you can also fly it manually. batteryu warning comes on a long time before there is not enough power to fire the solenoid. thats why we have redundant power supllies and controllersIf the battery is low it will not supply enough EMF for the oxygen solenoid
to open the valve and add life sustaining gas.
The handsets require being at ambient pressure and so HAVE to have the open cable. the handset problems are NOT due to this, but more due to the fact they are longer than the unit and so can get squashed, crushed and swung about. Normally while in air freightThe control handsets are secured to the canister via rubber hose, the
wiring
is run through this conduit down to the electronic handsets that are
monitoring the dynamics of the oxygen sensors. These rubber conduits enter
into the scrubber/O2 sensor compartment where it is humid and if not
perfectly sealed will allow condensation to migrate into the hose and wreak
havoc with the electronics in the handsets. Many electronic problems with
the handsets have involved this scenario.
The UK coroner was not aware of many facts, and had no advice from the manufacturer, The statistics were made up by someone claiming compensation and are not true. bit like most of your complaintsSee Mike Pizzios complaints on techdiver, see the archives of techdiver for
info on the Dsperation, see the UK coroner for what the bottom line on this
thing is. My take? "needs to be sold with a shovel".