Confined Spaces Question ANSI/ASSE Z117.1 - Daily Bump Test versus Calibration
Had a good question: "Tim can you tell me what Z117.1 states in regards to bump testing and calibration? Does Z117.1 require calibration prior to each use or per a preset schedule?"
Answer from Z117.1 is below. Please note that calibation versus a bump test has been one of the most debated question within the Z117 Confined Spaces Committee. Hopefully this will be of some assistance. And, I am not making this a commercial but the link to the Z117.1 standard is attached at the end!!!
Instrument Reliability Verification. Calibration of direct reading portable atmospheric testing instruments shall be conducted according to the manufacturer’s recommendations, or more often if necessary because of the instruments usage to ensure accuracy is maintained. At a minimum, a function check (e.g., bump test or a practice meeting the manufacturer’s recommendations) shall be done prior to each days use.
Full Span Calibration – Should be done in accordance with the manufacturer’s recommendations.