OK Joe, spill the beans, how do you get 92% O2 to register as 100% in the electronics? By pressurizing the head past 1 ata?
And how could I have gotten higher than 95% in places like Ambon Indonesia, where I'm pretty sure they only had a PSA system? My sensors were calibrated to good O2 back home and came back down to 20+% in air after cal with the O2 in Ambon, and the O2 there looked like about 97% on my 2dry at sea level. How common is an extra argon removal stage on PSA systems?
if its o2 by a psa system it CAN'T be greater than 95.5%, unless it has a second PSA system designed to remove argon then its 99.9%.. if you were reading higher than this the display is either not calibrated properly or the sensors are not linear..
look on rb world for an article Accurate po2 calibration..
a secondary argon removal system is not that common since it nearly doubles the cost of the system.. Its generally only used in aplications that demand the higher purity..