Question Flush O2 sensor with CO2 when not in use ?

Please register or login

Welcome to ScubaBoard, the world's largest scuba diving community. Registration is not required to read the forums, but we encourage you to join. Joining has its benefits and enables you to participate in the discussions.

Benefits of registering include

  • Ability to post and comment on topics and discussions.
  • A Free photo gallery to share your dive photos with the world.
  • You can make this box go away

Joining is quick and easy. Log in or Register now!

I have a sensor in an o2 analyser, probably around 500 tests over 7 ish years
I made a silicone seal for the cap which means I have a 100% seal out of use
Still using it, again today.

Think as you choose

Being sure, is better

 
There is no need to do that and it may cause the sensor to die. A while back I attended a workshop hosted by Analytical Industries engineers who emphatically told us not to try store sensors in inert or other gases.
 
Crap, thanks for sharing !
I've had my o2 analyser in a mylar bag purged with argon for 6 months ... It still registers 2/3% o2 in the bag but now I'll never trust it.
Rip I guess 😑
 
Crap, thanks for sharing !
I've had my o2 analyser in a mylar bag purged with argon for 6 months ... It still registers 2/3% o2 in the bag but now I'll never trust it.
Rip I guess 😑
Dead sensors won't get to 21%. If you can get the analyzer to read 21%, it's good. Gas analysis is all relative to a 21% calibration in air.

Going forward, store the analyzer open to the air in a cool/cold environment.
 