@tarponchik please read the Analox manual on calibration
https://www.analoxsensortechnology....2Analox_O2EII_RM-002-03_-_Calibration_Gas.pdf
it states specifically that you should calibrate with a gas at a higher fO2 than what you are trying to analyze.
They also recommend calibrating with 100% O2, and doing linearity checks with 50% O2
So since I just found that, and they agree with everything that I've written, are you now saying that the manufacturer of the analyzer in question doesn't understand calibration and how the oxygen cells function?