WT? If true, this incident has officially gone off into bizzaro land. Can you source the info or is it something you cannot yet disclose?
On the one hand, diving a fully assembled CCR that you didn't build is up there on the list of what I'd call insane: among other reasons, you cannot do the stereo check, which confirms the hearts of your one-way gas flow are actually working right, with the loop assembled on the unit. On the other hand, it would make this much, much more tragic than it already is...if someone else has to kick themselves for what happened.
If you were a novice rebreather diver and your instructor assembled it and were told this was O.K., would it still be insane to dive it?
Misplaced trust maybe, but not insane.
---------- Post added December 3rd, 2014 at 02:49 PM ----------
Husband and Wife purchase P2s and inital training. They complete the class together allegedly achieving certification.
This is pure speculation at this stage.
We do not know that, although it is possible.
Let me repeat what we know which at the moment has been made public (but there is more):
"It was so a Hollis.
It was sitting at the Health and Safety Laboratory Buxton Derby SK19 9JN ring them telephone number +44 1298 218000
SNIP
The rebreather had been put together such that the gas would not circulate in the required loop and CO2 would not be removed from the inhaled gas.
The inhale counter lung was positioned to the divers right hand side, instead of the left hand side.
The exhale counter lung was positioned to the divers left hand side, instead of the right hand side.
The one way inhale valve that should be positioned on the inhale side of the mouthpiece, to ensure the gas circulated the loop, was found butted against the one way exhale valve of the mouthpiece stopping any gas passing into the exhale counter lung on the divers right hand side.
The unit was flooded with fluid.
The cover for the scrubber bucket and head were missing, apart from a small piece attached to the fitting above the battery housing.
Two of the three oxygen sensors were reading 0.18 and 0.16 partial pressure of oxygen (ppO2) when exposed to air (0.21 ppO2), and when put into a 100% oxygen atmosphere (1.00ppO2) gave a reading of 0.80 oxygen. The third indicated 0.00 oxygen in air and 0.28 in 100% oxygen. Images received with the equipment taken by the Jersey Police indicate that the sensors appeared to be working correctly at the time of the incident. The subsequent reduction in oxygen sensor readings was due to the sensors being contaminated by the fluid within the rebreather damaging the oxygen sensors.
In UK terms this would be labelled as an accidental death, possibly misadventure, however, inquests in Jersey have narrative verdicts. The verdict in this case for Mrs Jillian Smith (age 41 years and 7 months) was as follows:*
That she died in the afternoon of Saturday 15th March 2014 at St Catherines Slipway, St. Martin, [Jersey] after having been found unconscious in the sea during an underwater dive at St Catherines Breakwater; the cause of death was asphyxia caused by obstruction of the airways by inhalation of gastric contents due to unconsciousness by hypercapnia; this occurred after failure of the incorrectly assembled rebreather diving equipment that she was using whilst underwater during her dive.
The inquest was informed that the manufacturer, Hollis, has now modified the mouthpiece assembly and re-issued it.
SNIP"